I have an ORP meter that looks practically the same as that, so assume it is from the same company. If it is anything like that, it is a single point adjustment only by a screw on the side. Did it come with an adjustment screwdriver? Assuming that is the case, you can do is buy a single calibration fluid closest to the value you expect to measure. You can then adjust it for that. The further you are away from that value, the greater the possible error is, but it would probably be close enough.